Re: Coax recomendations



One thing I've learned is that a (good) short circuit is a better termination for this kind of measurement than an open circuit. With an open circuit you get fringing which varies with frequency and cable diameter.

If you don't have a decent commercially made short circuit, you can do reasonbly well with a connector and several radial wires from the shell to the center conductor, or better yet a metal disk. Or you could put a male connector on a short piece of coax and short circuit the end of the cable by squeezing the braid into the center and soldering it to the center conductor, taking its length into account for the measurement of course. The object is to minimize the series inductance which would be created with a single-wire or similar connection.

Wes and Ian can probably point you to more possible causes for this measurement result -- they're the real experts on VNA measurements. I'd be very surprised if the VF really varies with frequency -- in theory, it can't, in an ideal coaxial cable.

Roy Lewallen, W7EL

Greg Ordy wrote:
Roy Lewallen, W7EL, wrote:


I measured 80% velocity factor, so if our respective
measurements are good, the velocity factor of your piece should be
around 72%.


Ok, my curiosity got the best of me, and I decided to
to measure the velocity factor. My VNA software has
a "distance to fault" feature, and I "worked backwards",
which is to say that I measured a length of the
BuryFlex with a tape measure (27' 2"), and adjusted
the VF on the distance to fault tool until I obtained the
same physical length. The far end of the cable was
terminated with an open circuit.

I happened to start with the measurement frequency set
to 1 MHz. Lo and behold, the VF needed to compute the
same physical length was 72%, as you suggested.

My own understanding is that VF should be constant with
respect to frequency, so I decided to vary the test frequency.
I should have left well enough alone.

I picked 1, 2, 4, 8, 16, and 32 MHz. For those frequencies,
I measured the following VFs:

1 MHz = 72%
2 MHz = 73%
4 MHz = 75%
8 MHz = 80%
16 MHz = 79%
32 MHz = 79%

I'm rounding the VF to integer values, since I don't think that any more accuracy
can be claimed in this setup.


Since that result was a little surprising, I grabbed some mini 8 (8X) that was nearby, about 51.25 feet. The published VF is 78%, and I measured the following, at the same test frequencies:

1 MHz = 78%
2 MHz = 78%
4 MHz = 79%
8 MHz = 79%
16 MHz = 80%
32 MHz = 80%

With this cable, the VF appeared much more constant across the
1 to 32 MHz range.

Is there an explanation that fits with my measurements?

Greg, W8WWV







.



Relevant Pages

  • Re: AC to DC conversion without using diodes and no external supply
    ... 200 mV peak is 140 mV rms and if that was unloaded voltage, ... voltage into a matched load would be 70 mV rms. ... If the original measurement was actually open circuit peak-to-peak, ...
    (sci.electronics.design)
  • Re: Sony KV-T25SZ8 TV problem
    ... > Capacitors are normally open circuit for DC, what do you mean by this? ... I'm wondering if my meter is stuffed. ... leg and then take the measurement. ...
    (sci.electronics.repair)
  • Re: regarding hybrid parameters
    ... And a reason that they are often more popular than other ... parameters is their relative ease of measurement. ... Certain ports ... are made short or open circuit to AC during testing which is ...
    (sci.electronics.basics)
  • Re: HP calibration resistor out of spec
    ... I've been in the biz 30 years, it was a 4 terminal measurement. ... I work in a cal lab and have for 30 ... Or just a coupla probes across the termination? ...
    (sci.electronics.design)
  • Re: Identifying Surplus Toroid Material
    ... I make a parallel tuned circuit with the inductor and an air variable capacitor. ... I've found that even mica capacitors often have a low enough Q to affect the measurement of reasonable Q inductors. ... I couple into and out of the parallel circuit with 1 pF capacitors, connecting one to a signal generator and the other to a 50 ohm termination and a scope. ... For ease in making measurements, I built a simple 3 dB switchable attenuator and put it in line with the signal generator, terminating the output in 50 ohms at the Q meter so the attenuator would work properly. ...
    (rec.radio.amateur.homebrew)